PS: I haven't looked closely at the MIT design. I did look closely over the Challenge requirements.<br>
<br>
There are some requirements in the Challenge that the MIT design might not meet. <br>
<br>
</div>
<div style="margin: 0px; font-size: 12pt; font-family: Calibri, Helvetica, sans-serif; color: black; background-color: rgb(255, 255, 255)">
Ventilators have basically push-air-in release-air-out. <br>
<br>
The push needs a trigger. It can be simple - N times per minute. It can be "when the pt breathes in, start pushing to help. Also start pushing if they don't after N seconds"<br>
<br>
The push needs an amount limit. It can be 0.7liters. It can be upto 40cmH2O. It can be both.<br>
<br>
The push needs a time limit - no more than N seconds, with at least M seconds between pushes<br>
<br>
The released air must go further out, not back along the input tube<br>
<br>
And the release-air-out also needs Positive End Expiration Pressure - restrict the output flow in some manner so the lungs keep at least upto 20cmH2O air pressure<br>
<br>
I don't know which of these the MIT design meets. I don't see an obvious set of triggers and the PEEP support may be rudimentary<br>
<br>
I can think of several easy ways to add these<br>
<br>
/Bevin</div>